Event Abstract

A systematic comparison of biodegradation behavior of absorbable metal: in standardized immersion, arterial bioreactor and in vivo study

  • 1 Southwest Jiaotong University, Dept. of Materials Science and Engineering, China
  • 2 North Carolina Agricultural & Technical State University, NSF Engineering Research Center for Revolutionizing Metallic Biomaterials, United States

Introduction: Absorbable metals have been widely tested in various in vitro environments to evaluate their biodegradation as vascular stent materials[1]. However, there exists a gap between in vivo and in vitro test results. A key step is to identify the relevant biochemical and biophysical microenvironments in test-systems. The purpose of this study was to establish an appropriate methodology for accurate and standardized determinations of degradation parameters in the pre- and post- endothelialization stage after stent implantation, which plays a vital part in predicting the fate of magnesium (Mg)-based stent.

Materials and Methods: As-drawn Mg wires of 99.9% purity with a length of 10 mm and a diameter of 250 μm (Goodfellow, USA) were used. In the standardized immersion test, Mg wires were immersed in DMEM solution according to the standard protocol ASTM-D1141-98[2]. The LumeGen bioreactor (TGT DynaGen® Series, USA) as a vascular bioreactor was chosen to  stimulate physiological aortal conditions (Fig. 1a). Each porcine abdominal aorta with a diameter of ~8 mm and a length of 5 cm was mounted into a chamber with a flow rate of 100 ml/min and a pulse pressure of 80-120 mmHg (Fig. 1b). In aortal in vivo test, rat abdominal aortas were used for implantation. Two Mg segments were symmetrically implanted into the lumen and wall of each aorta in the bioreactor or in vivo tests (Fig. 1c). Samples were analysed by X-ray computed tomography (CT).

Fig. 1.(a) Photograph of bioreactor. Mg wires in the wall and lumen of a porcine aorta in the chamber (b) and rat aorta in vivo (c).

Results and Discussion: One of two Mg wires exposed on the lumen and contacted the circulating medium to simulate a pre-endothelialization stage. The other Mg wire was embedded under the intima to simulate a post-endothelialization stage. The degradation product volumes, residual Mg volumes and average degradation rates were calculated utilizing the CT data. In Fig. 2, in vivo degradation was slower than in vitro degradation both in the standardized static and flow bioreactor conditions. In terms of the aortal bioreactor, flow convection on the lumen surface severely accelerated Mg degradation due to the increase of mass transfer, fluid shear stress and pulsatile stress, compared with low Mg degradation rate in the wall. In the aortal in vivo model, the degradation rate of Mg wire in the wall was faster than that in the lumen, and calcification was observed around Mg surrounding tissue. The degradation of Mg wire in the wall (diffusion limited region) was accelerated, which may result in the decrease of local pH from lysosome.

Fig. 2. Volume ratios of residual Mg and corrosion products in the static no aorta, and the aortal lumen and wall of the bioreactor and in vivo conditions for 3-day.

Conclusions: First-time investigation of degradation testing in the aortal bioreactor revealed that hydrodynamics plays a dominant role on the degradation. The established porcine aortal bioreactor and rat aortal in vivo model are expected to provide a better understanding of degradation mechanism of absorbable metallic stents.

NIH grant (1SC3GM113728-01); NSF-ERC for revolutionizing metallic biomaterials (NSF-0812348); National Natural Science Foundation of China under Grant (81330031); Thanks Prof. Deling Kong, Dr. Zhihong Wang and Ms. Yifan Wu from Nankai University for in vivo test.

References:
[1] ASTM-D1141-98: Standard Practice for the Preparation of Substitute Ocean Water Annual Book of ASTM, Philadelphia, Pennsylvania, USA (2004)
[2] Zheng Y F, Gu X N, Witte F. Biodegradable metals[J]. Materials Science and Engineering: R: Reports, 2014, 77: 1-34.

Keywords: in vivo, in vitro, Biodegradable metal, biodegredation

Conference: 10th World Biomaterials Congress, Montréal, Canada, 17 May - 22 May, 2016.

Presentation Type: New Frontier Oral

Topic: Biomaterials for cardiovascular applications, vascular grafts and embolic devices

Citation: Wang J, Liu L, Yun Y and Huang N (2016). A systematic comparison of biodegradation behavior of absorbable metal: in standardized immersion, arterial bioreactor and in vivo study. Front. Bioeng. Biotechnol. Conference Abstract: 10th World Biomaterials Congress. doi: 10.3389/conf.FBIOE.2016.01.01260

Copyright: The abstracts in this collection have not been subject to any Frontiers peer review or checks, and are not endorsed by Frontiers. They are made available through the Frontiers publishing platform as a service to conference organizers and presenters.

The copyright in the individual abstracts is owned by the author of each abstract or his/her employer unless otherwise stated.

Each abstract, as well as the collection of abstracts, are published under a Creative Commons CC-BY 4.0 (attribution) licence (https://creativecommons.org/licenses/by/4.0/) and may thus be reproduced, translated, adapted and be the subject of derivative works provided the authors and Frontiers are attributed.

For Frontiers’ terms and conditions please see https://www.frontiersin.org/legal/terms-and-conditions.

Received: 27 Mar 2016; Published Online: 30 Mar 2016.

* Correspondence:
Dr. Juan Wang, Southwest Jiaotong University, Dept. of Materials Science and Engineering, Chengdu, China, Email1
Dr. Lumei Liu, North Carolina Agricultural & Technical State University, NSF Engineering Research Center for Revolutionizing Metallic Biomaterials, Greensboro, United States, llumei@aggies.ncat.edu
Dr. Yeoheung Yun, North Carolina Agricultural & Technical State University, NSF Engineering Research Center for Revolutionizing Metallic Biomaterials, Greensboro, United States, yyun@ncat.edu
Dr. Nan Huang, Southwest Jiaotong University, Dept. of Materials Science and Engineering, Chengdu, China, huangnan1956@163.com